Column Overloading?
Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.
3 posts
Page 1 of 1
I am doing organic acid analysis and am seeing some changes in retention times for my external stdandard mix. The components are eluting about 0.1 mimutes sooner than previously. I am injecting anywhere from 400 to 11, 000 ppm, as that is what range my samples are in. I also am getting very good repeatability with the std but not with realtime samples. I am using polypropylene inserts, as I have only about 0.2 mls of each sample. Should I be diluting? Please help. My parameters are: Agilent Stablebond AQ column at 1.0 mls/min, at 30 C, 210 nm on the DAD, 20 mM pH 2.9 phosphate buffer, with 1% ACN. Injection volumes for my calibration are 1, 2, 3, 4, and 5 ul.
-
- tom jupille
- Site Admin
-
- Posts: 4978
- Joined: Wed Aug 11, 2004 4:55 pm
10,000 ppm = 10 ug/uL.
If I read your post correctly, that means that you are getting something over 50 micrograms at the high end (5 uL @ 11 ug/uL). That is definitely pushing things (you didn't say, but I'll assume a 150 x 4.6-mm column).
Are you seeing a variation in retention time with increasing concentration? And are you seeing increased tailing at increasing concentration? Both of those would point toward overload. Also, depends on retention time. A 1% shift from run to run might not be significant.
If I read your post correctly, that means that you are getting something over 50 micrograms at the high end (5 uL @ 11 ug/uL). That is definitely pushing things (you didn't say, but I'll assume a 150 x 4.6-mm column).
Are you seeing a variation in retention time with increasing concentration? And are you seeing increased tailing at increasing concentration? Both of those would point toward overload. Also, depends on retention time. A 1% shift from run to run might not be significant.
-- Tom Jupille
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374
-
- Posts: 12
- Joined: Fri Oct 08, 2004 4:39 pm
Yes, those are the column dimensions.The shift in retention times is minimal, about 1 %. There is a small amount of peak tailing. I have reduced the concentration of my standard by factor of 2 for the components that have the strongest signal and am rerunning it now.
Thanks for your advice.
Ricardo
Thanks for your advice.
Ricardo
3 posts
Page 1 of 1
Who is online
In total there are 31 users online :: 0 registered, 0 hidden and 31 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: No registered users and 31 guests
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: No registered users and 31 guests
Latest Blog Posts from Separation Science
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science