glycol analysis with DB-5

Posted: Fri Jul 27, 2018 12:55 pm
by control
Hi,

I need to analyze glycols with GC-FID, the compounds appear in EPA 8015 and it says that the extraction is by azeotropic distillation with direct injection of water samples. I have a DB-5, is this a column for this type of injection? i think not.
With this column, can i analyze glycols? If not, what other method or column do you recommend?

Thanks.

Re: glycol analysis with DB-5

Posted: Fri Jul 27, 2018 2:08 pm
by Steve Reimer
It has been done using a DB-5 but a wax column works much better. Limit the injection volume with water.
Simple injection of 0.5 uL should get you to low ppm for ethylene and propylene glycol.

Re: glycol analysis with DB-5

Posted: Fri Jul 27, 2018 2:13 pm
by control
So i can use a DB-5 with direct injection of water but with low volume, is that correct?

Thank you very much.

Re: glycol analysis with DB-5

Posted: Fri Jul 27, 2018 2:57 pm
by cjm
If your detector is an FID you're going to have a hard time keeping that flame lit after injecting water on a "5" column. Wax columns are what we use. We also found we need to be very aggressive with maintenance after injecting water.

Re: glycol analysis with DB-5

Posted: Sat Jul 28, 2018 2:01 pm
by rb6banjo
I'm betting your glycols won't give you nice peak shapes with the nonpolar DB-5 phase. The wax is your best bet.