building a calibration curve
Discussions about GC-MS, LC-MS, LC-FTIR, and other "coupled" analytical techniques.
8 posts
Page 1 of 1
Can somebody tell me how to have the calibration curve on my HPLC results. I have run an HPLC with calibration and changed in analysis method the result option in % normalization. But in the end I still dont have the calibration plot... Thanks!
-
- Posts: 586
- Joined: Mon Oct 04, 2004 3:00 am
There isn't much information to go on here, but if you have specified area normalization there is no calibaration for this in most data systems, it just gives area percent of the components detected. Please provide more specifics of you data system and application.
-
- Posts: 4
- Joined: Tue Jul 15, 2008 9:55 am
Thank you for your reply. Well I have specified according to the manual the %normalization and named the peaks. But actually there is no other information on how to get to the calibration results, only how to open the already existing plot. So I dont know how to get to the results where it gives me the calibration curve from the run...
-
- Posts: 621
- Joined: Tue Nov 27, 2007 1:20 pm
We still need more information !Thank you for your reply. Well I have specified according to the manual the %normalization and named the peaks. But actually there is no other information on how to get to the calibration results, only how to open the already existing plot. So I dont know how to get to the results where it gives me the calibration curve from the run...
What software system are you using? If we know this we can provide the help you need much more easily.
In most cases, %normalisation will only tell you what percentage area each peak has based on the total peak area of all integrated peaks in the chromatogram it will not perform a calibration for you.
Is there an option for "External Standards"? This type of option will allow you to enter the identity (vial No., concentration etc) of your calibration STDs in order to construct a calibration line for sample measurement.
Good judgment comes from bad experience, and a lot of that comes from bad judgment.
-
- Posts: 4
- Joined: Tue Jul 15, 2008 9:55 am
I am using Gilson HPLC with Unipoint V 1.71 software. If you could tell me a step by step process how to make a calibration curve it would be very helpfull. Thank you again very much..
-
- Posts: 621
- Joined: Tue Nov 27, 2007 1:20 pm
Sorry, I have no experience of this software system.I am using Gilson HPLC with Unipoint V 1.71 software. If you could tell me a step by step process how to make a calibration curve it would be very helpfull. Thank you again very much..
Do you have any User Manuals?
Have you tried using the on-line Help in the program?
Also, was there any form of Tutorial disk supplied with the software?
Any of these sources may help you to get started. Failing that, try contacting your Gilson sales rep ot their technical support.
Good judgment comes from bad experience, and a lot of that comes from bad judgment.
-
- Posts: 4
- Joined: Tue Jul 15, 2008 9:55 am
thank you very much for your reply.
I have checked in the manual there is no information about this. I am not the first one to use it so I'm afraid that there are no CD-s available with the manual. But there is also a possibility that the program is too old for a CD:) I will try to find an online information. Thank you again.
I have checked in the manual there is no information about this. I am not the first one to use it so I'm afraid that there are no CD-s available with the manual. But there is also a possibility that the program is too old for a CD:) I will try to find an online information. Thank you again.
-
- Posts: 2177
- Joined: Tue Aug 17, 2004 7:59 pm
Open up Excel
Put your standard concentrations in a column, put the corresponding areas in an adjacent column
Click and drag over both and make an X-Y scatter plot
select series
add trend line
format trend line to show equation
Once you figure out how to get the CDS to match this, you're home-free.
Put your standard concentrations in a column, put the corresponding areas in an adjacent column
Click and drag over both and make an X-Y scatter plot
select series
add trend line
format trend line to show equation
Once you figure out how to get the CDS to match this, you're home-free.
Thanks,
DR

DR

8 posts
Page 1 of 1
Who is online
In total there are 210 users online :: 0 registered, 0 hidden and 210 guests (based on users active over the past 5 minutes)
Most users ever online was 10230 on Thu Dec 04, 2025 5:56 am
Users browsing this forum: No registered users and 210 guests
Most users ever online was 10230 on Thu Dec 04, 2025 5:56 am
Users browsing this forum: No registered users and 210 guests
Latest Blog Posts from Separation Science
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science