Dear people on the forum,

I am writing with a strange phenomenon for which I am eager to find a solution, but cannot make it myself.

We have a Waters Premier QTOF which we use routinely for accurate mass measurements with positive mode ESI and for GAG measurements with negative mode nanoESI. MassLynx 4.1 is installed on the IBM PC provided by Waters.

The problem is that we do not have proper signal in negative nanoESI. The test sample is 0.5 ng/uL LeuEnk. In ESI 10 uL/min and in nanoESI 0.6 uL/min flow rate is used.
Working with ESI source we have a signal of ca. 1500 cps in positive mode and 300 in negative mode. (Even this is not the best ratio...)
But working in nanoESI there is a 2000-3000 cps signal in positive mode and there is no peak for LeuEnk in negative mode. Yet, there are some signals, but only under 350 m/z. Over 350 m/z there seems to be only very low electronic noise, nothing else.

This problem is repeatable, changing from negative to positive and vice versa we see always the same signals and intensity.

I am curious if somebody had this problem before and could provide a solution or some ideas?

Thank you in advance!

Gábor
The problem solved itself, but nobody knows how. I would like to understand the phenomenon, so if you happen to have some information, please contact me.
2 posts Page 1 of 1

Who is online

In total there are 12 users online :: 0 registered, 0 hidden and 12 guests (based on users active over the past 5 minutes)
Most users ever online was 599 on Tue Sep 18, 2018 9:27 am

Users browsing this forum: No registered users and 12 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry