Agilent 255 Nitrogen Chemiluminescence Detector
Posted: Mon Jun 30, 2008 3:35 pm
We have recently purchased the Agilent NCD to perform nitrosamines. The detector displays an excessively noisy baseline. The Agilent engineer performed an IQ and OQ and his runs were also noisy. We have tried increasing the quality of the oxygen, replacing the insert, replacing the burner chamber quartz tube, varying the oxygen flow to the burner and ozone to the reaction chamber, ensuring that the peltier cooler is performing at -20 and have inserted a CTR filter in the transfer line. We have tried both the 10 V and 1 volt output from the detector. We normally run our GCs at 10 V but also lowered this to 1 Volt and still suffered noisy baseline. Has anyone else used this detector, suffered the same problem and what did you do to resolve this? Oxygen flow set at controller at 5 ml/min and temperature for burner at 450. We are using HP5, 30 m, .32 column with .25 u film thickness. Insert is dual-tapered with glass wool. Pressure at reaction chamber is 2.2 torr.