Dealing with standard curve to determine concentration

Discussions about GC and other "gas phase" separation techniques.

2 posts Page 1 of 1

Hi Kankyo,

You could try including zero or force thorugh zero,but this will depend on the method. For low level evaluation using curves this is usually typical as the intercept value is 99.9% of the time a greater are value then the deg peak ,hence a negative number is obtained. Therefore y=mx ,not
y=mx+c. See what you get ,but be aware that 'including zero'and 'force through zero'are different !
hope this helps

Thanks

Andy

kankyo,

Two things. 1) You should run some sort of method detection limit study wherein you determine the minimum amount of analyte necessary for some signal to noise limit (say 10:1.) 2) You should not just rely on the numbers but you should look at the chromatogram as well. Is there a peak there or not? If so, is it significantly greater than the noise level around it? Don't just trust numbers churned out by data systems.

Best regards.
2 posts Page 1 of 1

Who is online

In total there are 101 users online :: 0 registered, 0 hidden and 101 guests (based on users active over the past 5 minutes)
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm

Users browsing this forum: No registered users and 101 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry