negative peaks HPLC

Discussions about gel permeation chromatography / gel filtration chromatography / size exclusion chromatography

3 posts Page 1 of 1
I have a problems with negative peaks. My systems is a gel permeation column with THF- methanol as solvent, I'm separating glicerol, triglicerides and ethyl esthers with an RI detector.

Can you be a bit more specific? Where do these peaks occur relative to your analytes? Are they always at the same retention times or do they move around?
-- Tom Jupille
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374

Generally speaking, if your samples contain compounds/solvents that have a refractive index lower than the refractive index of your mobile phase (i.e. TFA-methanol) then you will experience negative peaks when these compounds/solvents are eluted...