Chemstation and Macros
Posted: Tue Jun 24, 2008 5:24 pm
by lcrandall
Hi
I'm using the latest Chemstation and am having a problem getting macros to run.
In the run time checklist, I check "post run macros" and put in PUMPOFF. This is supposed to be a built-in macro. But, it doesn't run. My pump doesn't turn off at the end of the sequence.
Any suggestions?
Thanks!
Posted: Tue Jun 24, 2008 9:27 pm
by Consumer Products Guy
Try "Standby" instead of "Pumpall".
Posted: Wed Jun 25, 2008 1:26 pm
by lcrandall
Standby worked. Thanks!
I thought Pumpoff was supposed to be built in as well. I'm looking to keep my lamps on.
Thank you
Posted: Wed Jun 25, 2008 3:58 pm
by Consumer Products Guy
Yes, Pumpall "should" turn off the pump and leave the lamp on. I'd try it again. And re-boot all if necessary, and try again. We use both here, and both work, on Chemstations from A.06 to A.09.
Posted: Mon Jun 30, 2008 2:46 am
by mbicking
If you are running a sequence, you can also specify in the Sequence Parameters. There are several options.
Posted: Mon Jun 30, 2008 2:48 pm
by lcrandall
Thanks! The Pumpoff macro works when put into Sequence Parameters, but not into the run time checklist. As long as it works somewhere!
ChemStation Macro
Posted: Mon Aug 18, 2008 3:33 pm
by nimbi63
Actually it's a command, not a macro. The command to put in post run is "Pumpall Off". There is a macro in the core directory, Shutdown.mac, which can be modified to do the post run or post sequence settings required. As written it sends the Standby command which shuts off all lamps and pumps.
Posted: Mon Aug 18, 2008 3:57 pm
by lcrandall
Pumpall off worked once I restarted the computer.
Thanks!
Lisa
Posted: Sat Sep 12, 2009 11:01 pm
by bisnettrj2
I know this thread is old, but I'm hoping someone has an answer.
In the handheld Agilent controller, you have the option, upon finishing a sequence, to have one of several things happen - pump off, pump and lamp off, whole system off, etc. What I want to do is, upon finishing a sequence in Chemstation, shut off the pump and the lamp, but leave the column compartment heater on. Anyone know how to do this?