A plain silica HILIC column will act as a cation exchanger when pH is raised above 4, as the -OH groups forms negative silanol groups on the surface. Basic compounds then bind more strongly and more buffer salts are needed for elution.
The ZIC®-HILIC column is based on a bonded phase having zwitterionic functional groups, while a silica HILIC column is just plain silica.
The ZIC®-HILIC zwitterionic groups maintain their charged but overall neutral hydrophilic form, independent of pH-value. This opens for a unique selectivity since the pH of the mobile phase can be used to moderate the dissociation and retention of the analyte(s).
The ZIC®-HILIC phase allows creation of a stable aqueous layer on the stationary phase, and holds a lot of water at the surface. This makes the column very robust in both isocratic and gradient mode, and it allows aqueous samples to be injected (unless too large volumes are used). The column also has a very good lifetime.
A silica HILIC column has all the problems associated with straight phase chromatography; adsorption effects, peak band-spreading (for some compunds), and slow equilibration. It is furthermore sensitive to changes in water ratio in mobile phase, making it difficult to run gradients due to slow kinetics.
More information can be found at
www.sequant.com where you also can order, free of charge, the "Practical Guide to HILIC"
My suggestion is to add the zwitterionic ZIC®-HILIC column to your list.
/Patrik