I am sorry about my wrong expression of in-line degasser.I want to say that in-line degasser(installed in low-pressue mixed system) have the same ability of reduce bubble formation as high-pressure system.So,in-line degasser effect application of high-pressure system.Why are we worried about bubble formation and degassing in a high-pressure system. There are no bubbles formed before the pump, since nothing is mixed. There are no bubbles formed after mixin on the high-pressure side, since it is high pressure. Bubbles after the column in the detector cell can be avoided by putting a small restrictor capillary behind the detector, and many detectors have such a thing anyway. Or, in MS, you want the entire mobile phase to become a giant bubble...
So bubbles are not a problem in high-pressure mixing systems...