residual solvent method validation
Posted: Sat Jun 21, 2008 3:00 am
Hi, I have to validate a residual solvent method- i have also to design the protocol. I went to ICH guidelines and found that linearity should cover the range from the reporting level of the impurity to 120 % of the specification therefore, linearity should be performed from the QL to 120 % of the limit. So if the limit was 5000ppm then this would be from QL to 6000 PPM
Accuracy assessed on samples spiked with known amount of impurity at 3 levels/ 3 replicates. Therefore the drug substance should be spiked at 3 levels. (Maximum 120 %?)
Robustness should examine different columns and different temperature and flow rate?
Since residual solvents are volatile it seems pointless to look at solution stability for longer than 24 hours.
Please can someone let me know if I am going in the right direction
Thanks
Liv
Accuracy assessed on samples spiked with known amount of impurity at 3 levels/ 3 replicates. Therefore the drug substance should be spiked at 3 levels. (Maximum 120 %?)
Robustness should examine different columns and different temperature and flow rate?
Since residual solvents are volatile it seems pointless to look at solution stability for longer than 24 hours.
Please can someone let me know if I am going in the right direction
Thanks
Liv