Pore size is a normal parameter of column,what does it mean?
Here's a schematic diagram taken from our
Fundamentals of HPLC course that should help to explain. The particle size is 10 μm, the average pore size is 10 nm (= 100 Ã…).
Some manufacturer develop non-pore particle packing,what's its advantage?
The advantage is basically faster mass transfer (better efficiency at high flow rates) because analyte molecules do not need to diffuse in and out of the pores. The disadvantage is greatly decreased retention and loading capacity, because most of the surface area of a porous material is inside the pores.
Will it be a tendency of column parking?
First of all, it's not an "all-or-nothing" proposition; there are also "superficially porous" / "pellicular" / "porous layer bead" packings which have a non-porous core surrounded by a porous layer, so you really have to consider a continuum from totally porous to totally non-porous with all variations in between. There's nothing really "new" in the concept; materials of all these types have been around for over 30 years, and will probably continue to be around for the next 30!
Can it repalce present packing?
Per the previous paragraph, non-porous materials
are present packings. As with anything else, "the devil is in the details", and the tradeoffs among efficiency, retention, loading capacity, etc. must be considered in each case.