Here is the fragment about the method:
"LC-MS analysis of tris-hydroxamates in culture supernatants... ...An Agilent 1100 HPLC instrument
equipped with a binary pump and a diode array detector was used
for HPLC analysis. Samples were analysed on a Supelco Discovery
HSF5 column (150x4.6 mm, 5 mm i.d., column temp. 20 uC) and
eluted with 10 mM ammonium carbonate, pH 7.0 (solvent A)/
MeOH (solvent B) (10 : 90) at 1 ml/min for 10 min, followed by a
gradient to 100 : 0 A/B over 8 min, 10 min isocratic conditions at
100 : 0 A/B, a gradient to 10 : 90 A/B over 8 min and isocratic conditions
at 10 : 90 A/B for 4 min. Ferric-tris-hydroxamate complexes
were detected by monitoring A(435nm)."
The problem is that I am getting no separation!!! I can´t reproduce the cromatograms reported in the article. Everything elutes in the front...
First, I thought it was the mobile phase, so I changed the very hygroscopic carbonate by a new one, carfully checking the pH. Then I evaluated the column under other conditions and analytes (from the catalog) and I was able to achieve a separation... so, the column works.
Then I tried a 1:1 dilution of the sample with the mobile phase... still nothing. Everything still elutes in the front.
I ran out of ideas... please, enlighten me!!!