Gabapentin
Posted: Mon Jun 16, 2008 10:13 am
I tried to determine gabapentin using USP30-NF25 method for gabapentin and got this result:
Assay conditions:
Column packings : Phenomenex Luna C18(2) 5U 100Ã… 150x4.6 mm
Buffer solutions : 0.005 M monobasic ammonium phosphate + 0.015 M sodium perchlorate, adjusted with perchloric acid to a pH of 1.8
Mobile phase : Buffer solutions : Acetonitrile (80:20)
Diluent : 0.02 M monobasic ammonium phosphate, adjusted with phosphoric acid to a pH of 2.0.
Standard solutions : 6000 ppm Gabapentin in diluent
Column temperature : 45 C
Flow rate : 1.0 mL/min.
Detection : UV-Vis 215 nm.
The problem is that the tailing factor of gabapentin is >2.0
I already tried different ranges of mobile phase pH (from pH of 2.0 to pH of 7.5) but it didn't improve the tailing factor.
I also tried raising the column temperature up to 45 Celcius degree (from 40 degree) and adjust the mixture of the mobile phase, but still haven't achieve the best result yet.
What do you think I should try next to improve the tailing factor of the gapapentin peak?
I prefer not to use the derivatisation method and try to stick to the USP method since it's quite simple.
Assay conditions: