Negative Peak in Normal Phase HPLC during Method Transfer
Posted: Fri Jun 13, 2008 8:22 pm
Hi all,
Hope you can help me resolve this mystery.
I have developed a normal phase method to separate an amine with two chiral centers. Below is a brief summary of the conditions:
HPLC Column: Chiralpak AD-H chiral column, 4.6 mm x 250 mm, 5 µm
HPLC Column Temperature: 40oC
HPLC Mobile Phase: 80/20 (0.05%MSA in hexane/ethanol, 200 proof, v/v)
Flow Rate: 1 mL/min
Detector Wavelength: 210 nm
Injection Volume: 5 µL
Run Time: 15 min
Sample Diluent: Ethanol
Void Time: 3-4 min.
The method separates four isomersThe baseline and separation is good during the method development and qulification at our site using both old and brand new columns.
However when this method was transferred to an outside lab, a significant negative peak around 5.5 min, interfering with the main peak analysis. In addition, the separation of the last two isomers are not as good as the separation we observed. Discussion between the two labs revealed that there was some incompatible issues during system conversion, such as the reverse phased injector was purged with normal phase mobile phase without IPA wash in between. The problem was presumably corrected since, but withough improvement of the baseline and separation. It's also interesting to notice that the injection of mobile phase produces a small positive peak at the same retention time.
It's OK to switch the sample diluent to the mobile phase to eliminate the negative peak, but the separation still can't be reproduced. We are really puzzled since we never detect it before. Any clues will be greatly appreciated.
Flow Rate: 1 mL/min
Run Time: 15 min
Void Time: 3-4 min.