Specificity in Validation of a Limit Test
Posted: Thu Jun 12, 2008 3:54 pm
I'm working on a method validation of a reverse phase/ion-pair method method of analysis of EDTA in a natural carbohydrate polymer. The method converts the EDTA to the iron(iii) form and then it's a simple isocratic run with a water/methol with acetate buffer at pH 4 with tetrabutylammonium bromide.
Our specification for EDTA is < 0.1 %(w/w) and our three batches showed << 0.01%, so I'm revising the method to be a limit test where we run samples and compare to a standard prepared at 0.01% and report < 0.01%(w/w) as the concentration. Based on the ICH guideline, the table says to evaluate specificity and detection limit. Detection limit is straightforward.
For specificity, I don't have a true blank sample, where EDTA is undetectable, rather it is present in all samples. I can show resolution > 2 from another peak in the chromatogram. I can also show by spiking that there are no interferences that would cause a result of > 0.1% or even > 0.01%. Is peak purity necessary?
Thanks for any replies.
Our specification for EDTA is < 0.1 %(w/w) and our three batches showed << 0.01%, so I'm revising the method to be a limit test where we run samples and compare to a standard prepared at 0.01% and report < 0.01%(w/w) as the concentration. Based on the ICH guideline, the table says to evaluate specificity and detection limit. Detection limit is straightforward.
For specificity, I don't have a true blank sample, where EDTA is undetectable, rather it is present in all samples. I can show resolution > 2 from another peak in the chromatogram. I can also show by spiking that there are no interferences that would cause a result of > 0.1% or even > 0.01%. Is peak purity necessary?
Thanks for any replies.