Alcohol trace analysis with BSTFA and GC-MS
Posted: Fri May 25, 2018 9:44 am
by McCabe
Dear Experts,
I would like to measure the amount of alcohol (or -OH Groups) present in a rather complex sample, mainly consisting of various isomers of C21 hydrocarbons (aliphatic and aromatic) amongst other impurities.
My idea is basically to derivatize present alkohols with BSTFA+TMCS to Si(CH3)3 (m/z = 73) and to analyse the sample in SIM-mode, while scanning for the 73 ion. By that, I hope to obtain an "alcohol chromatogram" of my rather complex sample.
Is this a possible solution for my analysis or are there any better ways to achieve the desired results?
Thanks in advance and best regards
Re: Alcohol trace analysis with BSTFA and GC-MS
Posted: Fri May 25, 2018 2:56 pm
by Consumer Products Guy
Your derivatization idea should work fine; we often derivatized fatty alcohols with BSTFA to assay by GC and GCMS.
However, I'm not thrilled with using SIM, but don't want to argue against trying that. I think 73 ion will be contributed by other species in your matrix. How "trace" does your assay need to be?
That said: many trimethylsilyl derivatives will have a characteristic M - 15 ion using GCMS.
Posted: Tue May 29, 2018 4:45 pm
by MSCHemist
Would it be possible to use some sort of SPE phase to separate the oxygen containing analytes from the simple hydrocarbons?
Posted: Tue May 29, 2018 6:06 pm
by Rndirk
Are the alcohol(s) short chains? In that case, could you perhaps extract them from the oil with acetonitrile (or methanol, but this might be an analyte for you).