Help with RP peak tailing for protein

Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.

4 posts Page 1 of 1
Hello all,

I'm analyzing a 20 KDa protein by Reverse phase using a Jupiter C4 column. The problem is that the main peak has a persistent tail and contaminants actually co-elute during this tailing (confirmed by SDS-PAGE of fractionated run).

Image

Current conditions:
- Jupiter C4 column 250mm*4.6mm 5µm particle (new column)
- Buffer A: 5% ACN in water 0.1%TFA
- Buffer B: 95% ACN in water 0.1%TFA
- Gradient for 0% to 100% in 30 mns
- Temperature: 40ºC
-100µl injection
-Flow rate: 1ml/min
- Sample buffer: Buffer A

Already tried without success:
- Using propanol instead of ACN
- Higher column temperature
- C8 column
- slower gradient (more tailing)
- Different buffer sample
- Injecting 1/10 of the amount of protein.

All this attempts produced a peak with a similar shape.

I was inclined to try using TEA or acetic acid or formic acid instead of TFA.

Does anyone have any suggestions as what I could try to improve the peak shape and separation?

Many thanks for your attention,

Joana

Hi Joana,

Prepare the following eluents:
A – 0.1 M sodium phosphate pH 2.5 (adjusted before addition of ACN) add 5 % ACN.
B – 85 % ACN in water.
And try again.
Important! Remember to wash the column thoroughly (with 50 % ACN/water for 1 hour and then with the new eluents, running a couple of gradients) before the new conditions.

Best Regards
Learn Innovate and Share

Dancho Dikov

Hi Joana -

I recommend using a TFA gradient, but doing a 10-75%B.
If you're lucky - you'll notice a trend (which will help you for
further optimization).

If you look at Imtakt's RP protein data (using Intrada WP-RP) - it
usually starts with 10-20% organic:

Interferon: http://www.silvertonesciences.com/files/TI263E.pdf
Hemoglobin: http://www.silvertonesciences.com/files/TI272E.pdf
Bovine Serum (injected directly into the column):
http://www.silvertonesciences.com/files/TI274E.pdf

WP-RP is a novel design but it's reversed phase mode - so optimization principles should be
similar as for other reversed phase columns (though the results may be different).

Thanks Bryan and Danko for your quick replies.

I have already tried starting with a higher organic percent but the peak profile was still the same.

I will now try using the phosphate buffer at pH 2.5. I will report the results back later.

Thanks
4 posts Page 1 of 1

Who is online

In total there are 26 users online :: 4 registered, 0 hidden and 22 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Ahrefs [Bot], Google [Bot], Majestic-12 [Bot], Semrush [Bot] and 22 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry