Help.

Standard calibration curve not working .. only 2 standards at same level but intercept is much larger than the peak areas
ie zero amount has a peak response in calibration curve but not a calibration point
Yvalue - area , xvalue- amount, fit - linear- weighting - none

Y= -2.11E+003x + 3.61e+005 ?

X values
23.69
24.964
y value
311000
3308764

much appreciated
If you have only two calibration points at the same concentration, constructing a calibration curve by linear regression doesn't make much sense...
I'd use one-point calibration in this case.
Hi ,

Not exactly what I was asking but OK
What happens if method states mean 2 standards and use these to quantify by calibration curve
Not sure how or why, but once or twice in my experience, I changed from "Linear" to "Linear through Zero" in the processing method to get this sort of thing sorted. The original cal. curve looked like it had a negative slope(!).
Thanks,
DR
Image
ydna1977 wrote:
Hi ,

Not exactly what I was asking but OK
What happens if method states mean 2 standards and use these to quantify by calibration curve


It was exactly about what you asked :P . Your calibration model seems wrong.
If you have only two standards at the same concentration, constructing a calibration line by linear regression will give you nonsense. I don't work with Empower, but I'd suppose if you select "linear" as calibration model it will use linear regression.
If you calibrate at one concentration only (no matter how many replicates) you're doing a one-point calibration. This is indeed a linear calibration where the second calibration point is the origin 0/0. Either select "one-point calibration" (or whatever it's called in Empower), this will of course also be a calibration with a calibration line. If you still want to use linear regression (which mathematically doesn't make much sense in this case, but of course it's possible to do)you need to force the origin through zero.
5 posts Page 1 of 1

Who is online

In total there are 11 users online :: 0 registered, 0 hidden and 11 guests (based on users active over the past 5 minutes)
Most users ever online was 599 on Tue Sep 18, 2018 9:27 am

Users browsing this forum: No registered users and 11 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry