Rapid resolution HPLC pump

Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.

4 posts Page 1 of 1
Can anyone explain to me why an Agilent 1200 RRLC system sustains higher pressure than an 1100?
A.Neeley

I do not know what are the exact modifications but for sure they must have upgraded the injection valve and made general improvements in their autosampler...

Could you please clarify?.

Do you mean that for the same mobile phase flow, the system pressure is higher - if so that's probably due to using narrower bore components intended to reduce the 1200's internal volume.

If you meant, why can 1200 safely pump to a higher pressure, as noted by Kostas, that's due to the redesign necessary to compete with the Waters UPLC and similar high pressure instruments. Basically lower internal volumes, more accurate low flow settings, higher pumping pressure etc.

Bruce Hamilton

I wouldn't consider the Agilent 1200 (max pressure: 600 bar) an equal to the Waters UPLC (max pressure 1000 bar)...
4 posts Page 1 of 1

Who is online

In total there are 18 users online :: 0 registered, 0 hidden and 18 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: No registered users and 18 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry