GC-ECD Low response

Discussions about GC and other "gas phase" separation techniques.

3 posts Page 1 of 1
Hi guys, I am just wondering if you all have any idea to why the GC-ECD detector Agilent brand has a much lower response (five times less) response after I did the thermal ECD cleaning ?

Some of the peaks totally disappeared. This is not only on one method but two different methods so I can eliminate out column problem . We changed the liner and septum and trimmed the column too. But the response was still so Low.

We checked the detector part and noticed the detector screw a bit loose so tightened using a wrench . Is that why it have affected the response?

Thank you for the suggestions and I really appreciate it!

Regards
Hts123
Hi,
What is the signal of the ECD when the GC is in the idle state?

Also what model are you using?

This could give you an indication of contamination or an issue with the ECD cell.
Thank you for the responses.
I think the way I installed the column into the ecd detector caused the low response. I did not know that there must be a certain length of the column measured before installing into the detector. After following the guidelines as mentioned below in the link, I noticed a lower output for the detector and also a more stable baseline and defined peaks.

https://www.agilent.com/cs/library/quic ... erence.pdf

Thank you for the help.

Regards,
hts123
3 posts Page 1 of 1

Who is online

In total there are 16 users online :: 0 registered, 0 hidden and 16 guests (based on users active over the past 5 minutes)
Most users ever online was 599 on Tue Sep 18, 2018 9:27 am

Users browsing this forum: No registered users and 16 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry