GC-ECD Low response

Discussions about GC and other "gas phase" separation techniques.

2 posts Page 1 of 1
Hi guys, I am just wondering if you all have any idea to why the GC-ECD detector Agilent brand has a much lower response (five times less) response after I did the thermal ECD cleaning ?

Some of the peaks totally disappeared. This is not only on one method but two different methods so I can eliminate out column problem . We changed the liner and septum and trimmed the column too. But the response was still so Low.

We checked the detector part and noticed the detector screw a bit loose so tightened using a wrench . Is that why it have affected the response?

Thank you for the suggestions and I really appreciate it!

Regards
Hts123
Hi,
What is the signal of the ECD when the GC is in the idle state?

Also what model are you using?

This could give you an indication of contamination or an issue with the ECD cell.