Non-linearity When Working on the Slope of the UV Spectra
Posted: Wed May 16, 2018 11:20 pm
by adam
Hello
This seems like a basic question, but I cannot quite figure out what is causing it.
When doing HPLC with a UV/Vis detector, we often set the wavelength to a flat part of the curve (often the maxima). However, there are cases where we intentionally chose a less sensitive wavelength (for example, when we have multiple peaks and one has too much absorbance, and so would be off scale). But what I have found is that when using a detection wavelength that is on the slope of the spectra - and not on a flat part of the spectra - it sometimes causes a non-linearity in the response, or results in a significant intercept.
Now you may be thinking: There could be some slight drift in the position of the diffraction grating. And certainly that could cause it. But we are seeing this phenomenon with a Diode Array detector and, therefore, this cannot be the explanation.
Thanks in advance for any assistance. I would like to get a handle on why this is happening.
Sincerely
Adam
Re: Non-linearity When Working on the Slope of the UV Spectra
Posted: Fri May 18, 2018 9:18 am
by Khanom
Dear Adam,
What bandwidth you are using on your diode array detector.
Posted: Fri May 18, 2018 10:56 am
by HPLC chemist
Be sure you have 'Turned Off' the reference wavelength (minimum wavelength absorption). A slight drift in this area can cause non-linearity.
Posted: Fri May 18, 2018 12:32 pm
by lmh
could it also be that you're only doing this with things that give very high absorbances, to get them into a "linear range", but maybe the linear range isn't as high as it should be, and therefore all of these things are coming out non-linear, while the things that are present at low abundance (and therefore that you measure at a peak-apex in the spectrum) are actually a lower absorbance and therefore have better linearity?
Posted: Fri May 18, 2018 2:39 pm
by Consumer Products Guy
Wouldn't it be easier to dilute before injecting or to inject smaller volume?
We never used a "slope" wavelength because the absorbance at the maxima was too large, but occasionally did when there was a matrix interference at maxima wavelength but none/considerably less on the slope. Just saying that we didn't do that, don't know if that part is relevant to you.
Posted: Tue May 22, 2018 7:39 am
by bunnahabhain
Maybe what you are seeing is related to the fact that bands in UV spectra are broadening when concentration increases.