I am trying to troubleshoot a problem I faced when analyzing a water solution of bupranolol 15 ng/ml (system suitability test). I had a high signal intensity in the first injection (A=208718630), but in the second (A= 6081794 ) and subsequent injections the intensity dramatically decreased. After 15-16 hours (on the next day) the situation repeated (the first injection gave a high intensity, whereas subsequent injections yielded a much lower intensity).
I was advised to change rotor seals in 10-port valve and 6-port valve (autosampler valve). But now I do not have new rotor seals. I also try to use a ceramic emitter (liquid junction) instead of a stainless emitter but it does not help. I also check the stability of QExactive with the HESI-II source by infusion of calibration solution directly from the syringe pump (direct infusion) at 2 microliter/min for 4 hours. The signal is constant.
Has anyone faced a similar problem? Any suggestions or tips would be appreciated.
The problem was solved by cleaning the ion source (ion transfer capillary, exit lens, S-lens, octopole, etc), quadrupole and C-trap. As I understand, loss of the signal over time was caused by the charging effect. My thanks to Thermo Service for help.
2 posts Page 1 of 1

Who is online

In total there are 13 users online :: 0 registered, 0 hidden and 13 guests (based on users active over the past 5 minutes)
Most users ever online was 599 on Tue Sep 18, 2018 9:27 am

Users browsing this forum: No registered users and 13 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry