System suitability test FAILED
Posted: Fri May 11, 2018 10:23 am
by dhia djaadi
Hello, i launched the USP technic dedicated to Azithromycin tablets organic impurities dosage. The System suitability solution were injected six times , unfortunately the RSD for Area were higher than the 2.0% mentionned (3.6% for azithromycin related compound F , and 13.4% for desosaminylazithromycin) . The case is that the fifth injection were shifted to the right and weren't stackable with the others . Any suspected reason for that ?
Re: System suitability test FAILED
Posted: Sat May 12, 2018 8:34 pm
by tom jupille
I can think of many possible reasons, but the fact that you saw a retention time change on one of the injections suggests a flow rate problem. One possibility is that the one bad injection is the cause of the high RSD (what does the RSD look like for the remaining 5 injections?).
Also, is there a trend in the response (in other words, is the high RSD coming from a gradually increasing or decreasing response rather than random variation)?
Have you seen acceptable RSD with other methods run on this system ?
Posted: Wed May 16, 2018 1:54 pm
by dhia djaadi
Thank you for answering .
The retention time increased from injection 1 to injection 6 , and the RSD for this method is mentionned in the monograph =< 2%.
This is the summary of the technic results
And this is the graph of the sixth injection .
Posted: Wed May 16, 2018 11:59 pm
by tom jupille
Unfortunately, the images you linked to are not accessible. Error message was:
"403. That’s an error.
Your client does not have permission to get URL"
Posted: Sun May 20, 2018 1:19 pm
by dhia djaadi
Sorry it was my mistake , here are links to access to the pictures on google drive
https://drive.google.com/open?id=0BwJxH ... 01YWhueFB3
https://drive.google.com/open?id=0BwJxH ... xfRjVxU19n
and sorry again for being late
Posted: Thu Jul 05, 2018 3:40 am
by dap
Hello. As far as I see it seems to be an integration issue. Try to tweak integration options and reprocess your data or try manual integration.