Best approaches to gradient with ion pairing agents (odsa)

Posted: Sun Apr 29, 2018 1:59 am
by Geof235
Hi.

Sorry in advance for the long winded explanation...

I'm trying to separate about 12 known impurities from a two component drug product. After some method development I decided on developing a gradient method using ion pairing agents. Below is the full chromatographic condiitions.
Everything started going well during validation, multiple columns and multiple systems were reproducible. Then all of a sudden we started injecting LOQand resolution and we were getting a TON of noise making integration difficult.

Now I have a very concentrated solution of one active and reasonable concentration of the other (65,000ug/ml compound a, 500ug/ml compound b). Consequently, the placebo load as you could imagine is quite high.

I've heard that making gradients with odsa Is not recommended however I didn't have much of a choice. I needed enhanced selectivity and the main component and impurities are ketones and according to what I've found online exhibit extreme fronting due to gem diol formation and hemiketal formation.

I also noticed that when sample was injected prior to system suitability that the noise seemed to reduce. I tried making 3 quick injections of selectivity soliton followed by the normal gradient and then followed by 1 more injection of selectivity and then start into my system suit. When I did this however I am only seeing solvent front and component a, then no other peaks. Noise seems to have been reduced but again I don't see anything except solvent front and gradient baseline shift

1. Out of general curiosity, was my approach towards making a gradient method good? The idea was to as accurately as I could get the SAME concentration of ODSA in mpa and mpb? I was noticing when I first was playing around with this that I was getting some baseline noise but it was tolerable but then recently something happened and now I peaks are masked by an extremely noisy baseline.

2. In my attempt to "condition" the column, could I have gone overboard and that's why I'm seeing no peaks (other than the 65,000ug/ml component)?

3. As a side note, during the development we had also observed that different manufacturers of ODSA gave slightly different selectivity. This was confirmed by a colleague of mine and I was wondering if anyone had any ideas about why this is? We ended up during validation to stick with the same manufacturer of ODSA, and had opened a few different bottles from that same manifacturer but then we opened up smoke new one and we started observing the baseline noises.

Stock buffer solution: 10grams octanesulfonic acid, 2.5 grams potassium phosphate mono basic , 4ml tea per liter ph 3.0

Mobile phase a: 460ml buffer: 500ml water: 40ml acn
Mobile phase b: 460ml buffer:40ml water:500ml acn

Chromatographic conditions:
Injection volume 10ul
Flow:1.5ml/min
Temperature:50c
Column: Luna c18(2) 100x4.6, 5um
Wavelength: 235nm

Gradient:
0->15 minutes 100%a
15->16minutes 93:7
16->25minutes hold 93:7
25->45 50:50
Return to initial

Re: Best approaches to gradient with ion pairing agents (odsa)

Posted: Sun Apr 29, 2018 11:13 am
by HPLC chemist
I don't see much that needs correction. The only thing that could be a problem is that the formulation could build up on the column under these chromatographic conditions and then dumped into the injection. Try reversing and washing the column with 50% IPA.

Then add a washing procedure with 50% IPA after every injection. Then re-equilibrate the column with Mobile A before the next injection. Always use HPLC solvents.

Re: Best approaches to gradient with ion pairing agents (odsa)

Posted: Mon Apr 30, 2018 11:43 am
by hossam kamal
hi Geof235:
can you attach some chromatograms for illustration , i am sorry , but i think chromatograms will explain the problem more easily and so we can think with you

Re: Best approaches to gradient with ion pairing agents (odsa)

Posted: Mon Apr 30, 2018 2:44 pm
by Vlad Orlovsky
In my opinion, the best approach for gradient with IP reagent is not to use ion-pairing reagent and explore mixed-mode chromatography where IP is attached to the surface of silica gel :lol:

No need to dedicate column to be used with IP reagent, much shorter equlibration and full compatibility with LC/MS and prep chromatography.

Please contact me if you want my help (http://www.helixchrom.com)