For many months, we have had trouble with unstable retention times (Tr). The Tr RSD can sometimes be over 1% and my peak tables become useless. The vendor has already tested, then replaced the whole pump (with built-in degasser and proportioning valve), but there was no improvement. New columns, freshly-made buffer: no improvement! We have tested 3 different standards (ethylene glycol, ethanol, and acetone) to make sure this isn't related to our normal standard (EtOH). The filters (in eluent bottle, and in pre-column cartridge) have been changed several times. In some experiments, the pressure has increased dramatically between injections (e.g. from 20 to 35 bar), but this has never caused this type of Tr variations before. We have also checked the flow rate with a volumetric flask: it was only within 95% of the required rate (0.95 vs. 1 mL/min), but the vendor insured us that this was normal. Is it possible that the autosampler is causing all this trouble, by e.g. delaying injection?
Hardware: Dionex UltiMate 3000 (quaternary pump with built-in degasser and proportioning valve; autosampler; oven with column selection valve; 4-wavelength UV detector; RI detector)
Columns (HPSEC): Shodex OHpak SB-805 and -804, in series, with guard column
Method: isocratic, 1 mL/min, 30 min
Eluent: phosphate buffer, 10 mM, pH 7.2; with NaCl 0.15 M