Can I know what is the effect of internal diameter of column on its sensitivity/column performance??
thanks
Advertisement
Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.
I’m not quite sure I understand your reasoning, but the fronting on the first chromatogram is plain to see and I’d certainly call it distortion.think the analyte in the first chromatogram is well retained.
So - there wasn't any distortion when the i.d. was reduced from
4.6mm to 2.0mm
Sorry - 4.6mm because extra column volume effects are minimizedHi Danko -
Thank you for your discussion. I think we all can agree that there
are benefits to using smaller column i.d. We have lots of LC-MS users using our 2mm i.d. - they're detectors work best at lower
flow rates - and our 3um particle size (2mm id) works very well at
lower flow rates. (Some of the newer MS detectors can handle
higher flow rates - so some have switched to use our 4.6mm i.d.).
The point I wanted to make was that decreasing i.d. will do nothing to increase separation. For example: If an HPLC user wants
to go from 150x4.6mm down to 150x3.0mm to increase resolution - we advise against it. It will do nothing to increase alpha
(in other words, just because 2 bands are more concentrated
doesn't mean their migration rates will change).
For conventional HPLC systems - the 4.6mm id will give you the best opportunity for separation.