- Posts: 3210
- Joined: Thu Sep 02, 2004 7:28 pm
You must have a very SHORT column with a VERY THIN film. It is not the phase of choice. FFAP will STRONGLY retain cholesterol and you will have to use the MAX temperature for a very LONG time. I hope it will elute.
The best separation of cholesterol and impurities is on a capillary OV-17 (SPB-2250) and will take a good half an hour to elute for best separation. TMS derivative is suggested.
You may retire before cholesterol elutes from a FFAP column. Do not use a TMS derivative for this column.
GOOD LUCK.
Rod
The best separation of cholesterol and impurities is on a capillary OV-17 (SPB-2250) and will take a good half an hour to elute for best separation. TMS derivative is suggested.
You may retire before cholesterol elutes from a FFAP column. Do not use a TMS derivative for this column.
GOOD LUCK.
Rod
-
- Posts: 6
- Joined: Wed May 14, 2008 11:59 am
3 posts
Page 1 of 1
Who is online
In total there are 105 users online :: 1 registered, 0 hidden and 104 guests (based on users active over the past 5 minutes)
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm
Users browsing this forum: dlntx9 and 104 guests
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm
Users browsing this forum: dlntx9 and 104 guests
Latest Blog Posts from Separation Science
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science