Tablet Analysis
Posted: Tue May 20, 2008 7:14 pm
by Chris
I am performing some HPLC training on tablets and need a method that is going to be as quick as possible. The following parameters are prefferable:
Active = non pharmaceutical (e.g. just a Vit C tablet or something similar)
Tablets can be dissolved straight into solvent (i.e. no grinding)
Reverse HPLC analysis (pref C18) with UV detection
Short run-time (less than 10 min per sample)
Anyone got any good ideas?
Posted: Tue May 20, 2008 10:51 pm
by tom jupille
If you don't mind OTC pharmaceuticals, use APC (aspirin / phenacetin / caffeine) tablets (Exedrin or your local drugstore house brand equivalent). Should give 4 peaks (the three listed plus benzoic acid).
Posted: Fri May 23, 2008 12:34 am
by syx
You will get problem when using vit C since this substance degraded rapidly in aqueous solution and difficult to retain in general RP18 column.
Posted: Fri May 23, 2008 7:48 am
by rp18
We're using paracetamol/aspirin/caffeine tablets (similar to Tom's suggestion) in our practical HPLC course (university).
Finished in 7 min, UV at 280 nm, RP 18 column with MeOH/water/acetic acid (400/600/1). Dissolving takes about 1 hour according to the Pharm. Eur.