Currently running a basic FID test mix (C14,C15,C16) and looking to achieve area uV/sec >200,000 for the instrument to pass operational qualification stage.

When using a manual injection this is great as consistently getting >600,000 however when I change to using the autosampler the areas significantly reduce to anywhere between 25,000 and 250,000 (usually at the lower end).

There doesn't seem to be a particular pattern to when spec passes. I have performed full system maintenance including change of autosampler and syringes but the same issue occurs.

Any ideas?

Thanks in advance.

Re: Varied Areas using Autosampler

Posted: Thu Apr 12, 2018 7:21 am
by Poly Styrene
Is there enough solvent in your vials, i.e. can the syringe withdraw a consistent amount of sample? And what injection volume are you using?

Re: Varied Areas using Autosampler

Posted: Thu Apr 12, 2018 7:30 am
by pumpedupchemist
There is 1mL of sample in the vial and can easily withdraw enough; I've even watched the autosampler to ensure no air is getting into the syringe. Injection volume is 1uL

Re: Varied Areas using Autosampler

Posted: Thu Apr 12, 2018 9:14 am
by pumpedupchemist
Third time lucky with another autosampler plus a user defined programme of a pre-injection delay seems to resolve the issue.

Re: Varied Areas using Autosampler

Posted: Thu Apr 12, 2018 4:06 pm
by James_Ball
Third time lucky with another autosampler plus a user defined programme of a pre-injection delay seems to resolve the issue.
That is sometimes the resolution to the problem if the solvent is not pulling up into the syringe quickly. Agilent I know has a box in the autosampler setup for Viscosity Delay, newer software also has plunger speed that can be set slower to reduce problems.