Bubbles in VOC samples

Posted: Wed Apr 11, 2018 2:05 pm
by Bigbear
Something that has been in the back of my mind for 20 years or so.
How do bubbles affect results (if any) when using a needle type autosampler that moves sample via purge gas pressure.
We accept samples with bubbles <3mm and run them without qualification.
Our sampler (Aquatec 70) over pressurizes the vial durring sample transfer to the P&T, so I would think reasonable bubbles should not make a difference.
I never got around to doing a study on this question.
Any of you done a study? How do you deal with bubbles in a sample?

Thanks
Bear

Re: Bubbles in VOC samples

Posted: Wed Apr 11, 2018 5:50 pm
by James_Ball
As far as sample acceptance criteria, I believe you have to keep it as ,3mm bubble, since overpressuring the sample for a short period of time to transfer to the purge and trap would not be long enough to force any analytes that have moved into the headspace/bubble back into solution. The bubble rule is mainly for loss of analytes during storage.

Re: Bubbles in VOC samples

Posted: Wed Apr 25, 2018 4:06 am
by mckrause
It is compound dependent, as you would expect. You (for obvious reasons) lose the gases pretty quickly if you have bubbles. You also, for whatever reason, lose perchloroethylene. However, I kept some vials of plain ol' groundwater with about 5 mm of headspace sitting around the lab (not refrigerated) for 6 months that I had spiked with BTEX compounds and we were still getting recovery when we quit playing with them, so the loss of water solubles is pretty low.