Dear All
I am presently setting up a GC method to analyse SCFA levels (acetic, propionic and butyric acid) in plasma/serum samples (which will be deproteinised prior to injection). I started off following a procedure that was previously set up in other labs, however I have already made a few changes to the method and still have a number of issues with it.
We are using a Stabilwax-DA column (30m x 0.53mm id, 0.5um df) from Restek, which is similar to the Agilent HP-FFAP columns, and does not require you to derivatise the acids. I have been running standards at the levels you would expect in plasma (1, 2, 3, 5, 8, 10, 20, 50uM propionic and butyric acids; 5, 10, 15, 20, 30, 50, 100, 200 uM acetic acid) made up with DD water. The standards include ~1% w/v formic acid (which was advised in the procedure we are using). Our internal standard is isovaleric acid.
While results for butyrate and acetate have been good, my problem is that the formic acid peak is running into the propionic acid peak, making accurate quantification difficult - a problem has not been solved by changing the oven T ramping.
So... my questions are:
1. Is the formic acid necessary, or could I use an alternate acid to acidify my samples? I had considered HCl, but having read other posts on this forum I'm not so sure. Any suggestions would be most welcome.
2. Would the Stabilwax-DA able to tolerate samples that had not been acidified (apart from the SCFAs present)? In other words could it be problematic/damage the column by not adding formic acid or any other acid at all?
Any other suggestions/comments would be most welcome, as I am very new to GC analysis!