SPME-GC-MS problems, help
Posted: Fri May 16, 2008 1:10 am
To anyone with experience using SPME-GC-MS,
I have a started to use SPME to analyze aromatic Gasoline Range Organics (BTEX,) and MTBE in Groundwater, however, I'm running into problems. Reason for picking SPME are obvious fast, cheaper, not as complicated as purge and trap and results are comparable to P&T. The fiber I'm using is a supelco 85 um Carbowaxen/PDMS, which was used in a paper that I'm following published in the: Journal of Chromatography A, Volume 1021, Issues 1-2, 22 December 2003, Pages 157-164
Fu Fang, Chia-Swee Hong, Shaogang Chu, Wenpeng Kou, Anthony Bucciferro.
The GC injector is equipped with the 0.75 mm ID inlet liner and I'm using the 11mm LB-2 septa. The capillary is Supelco VOCOL (10m x 0.2 mm x 1.2 um). The problem that I'm running into is that I'm getting severe band broadening and tailing for most of my peaks (some peaks even appear to have "Christmas Tree" shape), and I'm also having problems extracting MTBE. This has occurred with my 250 ug/L standard. I conditioned the fiber according Supelco. I believe I'm having problems focusing the analytes onto the column. The conditions are the following:
SPME-
1.Place 10mL of sample in 15 mL SPME Vial.
2. Add NaCl and stir bar, incubate and stir at 60 C
3. Insert SPME man holder and press down to allow fiber to absorb HS, extract for 10 min
4. Desorb in GC injection port for 2 min at 220 C
GC/MS
1. 35C for .5 min, ramp to 200 C at 25C/min hold at 200C for 1 min.
Before moving any forward, I descided to inject 1 uL of my PDS in the instrument with a split ratio of 100:1. All the peaks appeared sharp, as they should be, and are eluting about 0.5 min later. To make matters even worse is that I'm doing this all on brand new Agilent 5975c GC-MS, and I'm under a deadline to get this method developed and validated. Could the fiber be damaged already? Should I try starting at 50 ug/L?
I have a started to use SPME to analyze aromatic Gasoline Range Organics (BTEX,) and MTBE in Groundwater, however, I'm running into problems. Reason for picking SPME are obvious fast, cheaper, not as complicated as purge and trap and results are comparable to P&T. The fiber I'm using is a supelco 85 um Carbowaxen/PDMS, which was used in a paper that I'm following published in the: Journal of Chromatography A, Volume 1021, Issues 1-2, 22 December 2003, Pages 157-164
Fu Fang, Chia-Swee Hong, Shaogang Chu, Wenpeng Kou, Anthony Bucciferro.
The GC injector is equipped with the 0.75 mm ID inlet liner and I'm using the 11mm LB-2 septa. The capillary is Supelco VOCOL (10m x 0.2 mm x 1.2 um). The problem that I'm running into is that I'm getting severe band broadening and tailing for most of my peaks (some peaks even appear to have "Christmas Tree" shape), and I'm also having problems extracting MTBE. This has occurred with my 250 ug/L standard. I conditioned the fiber according Supelco. I believe I'm having problems focusing the analytes onto the column. The conditions are the following:
SPME-
1.Place 10mL of sample in 15 mL SPME Vial.
2. Add NaCl and stir bar, incubate and stir at 60 C
3. Insert SPME man holder and press down to allow fiber to absorb HS, extract for 10 min
4. Desorb in GC injection port for 2 min at 220 C
GC/MS
1. 35C for .5 min, ramp to 200 C at 25C/min hold at 200C for 1 min.
Before moving any forward, I descided to inject 1 uL of my PDS in the instrument with a split ratio of 100:1. All the peaks appeared sharp, as they should be, and are eluting about 0.5 min later. To make matters even worse is that I'm doing this all on brand new Agilent 5975c GC-MS, and I'm under a deadline to get this method developed and validated. Could the fiber be damaged already? Should I try starting at 50 ug/L?