You have a good column but the wrong one for the mix you are trying to separate.
You should one of two columns:
A molecular sieve 13X or a Carboxen 1000.
You can use either helium or argon as the carrier gas. Hydrogen sensitivity is poor with helium.
The type of packed column is determined by the concentrations you are trying to measure and the type of detector you are using.
If you have to measure below 1000 ppm of hydrogen then a micropacked is suggested with a TCD.
The chromatogram below shows fixed gases (H2, O2/N2, CO, Methane, CO2) at 120°C using a 6 meter SS 2mm ID 60/80 mesh Carboxen 1000 column using He carrier at 30cc/min. (45 psi) 4% each with a nitrogen balance (200µL) Note the low response of hydrogen with He carrier.
If you email me I will be glad to offer suggestions.