Gas saver?

Discussions about GC and other "gas phase" separation techniques.

3 posts Page 1 of 1
I go through my compressed air quicker than my other gases, is there a way to save gas without compromising my chromatography? The instruments are set up at 400ml/min for Air flow.

Are you using bottled air for your FID?

Oh MY !

Most places have a dedicated air pump with scrubbers to remove hydrocarbons and any organic comtamination from the air.

Atmospheric air is processed to be clean for use with an FID.

I cannot imagine the expense of using bottled air, nor do I ever want to.

You should really contact your gas supplier or one of the many chromatography businesses about obtaining another solution to your air requirements. (I work for such a company)

best wishes,

Rod

nabearup,

Your FID will need the 400 ml of air to run properly. The only other way to conserve air is to turn you FID's off when they are not in use and turn off the gas supplies to the detector. You can leave the detector hot with carrier and make-up gas an then just turn on the FID when you need it.

Best regards.
3 posts Page 1 of 1

Who is online

In total there are 94 users online :: 1 registered, 0 hidden and 93 guests (based on users active over the past 5 minutes)
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm

Users browsing this forum: dlntx9 and 93 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry