lpenney » Tue May 06, 2008 12:40 pm
I would like to try and answer some of the questions that have been asked. I'm sorry if I'm not answering them satisfactorily, but I'm not an expert in GC-MS!
- the new column is a Zebron ZB-50 from Phenomenex, 30m x 0.25 mm ID, 0.25 um film thickness
- we are using constant pressure mode with a starting flow of 1.1 mL/min
- we do an autotune on a regular basis (I'm not sure what the difference is between an autotune and a mass or resolution calibration). There have been no signs from the autotune that the source needs a cleaning (ie from the last autotune, the EM volts were approximately 1700, the abundance of the peak at 502 m/z was 64000, etc).
- I'm not sure when the multiplier was last replaced, if ever (I just started in this lab about 8 months ago).
Does any of this help to explain a loss in sensitivity? Is there something obvious that we're missing?
Thanks again for the help,
Lisa