Benzyl alcohol keeps failing in my initial calibration
Discussions about GC-MS, LC-MS, LC-FTIR, and other "coupled" analytical techniques.
2 posts
Page 1 of 1
We analyze 8270s on a GC/MS. All compounds pass perfect on averaged curves for our initial calibration except for benzyl alcohol. Any suggestions as to what may cause this?
-
- Posts: 5
- Joined: Mon Sep 08, 2008 7:47 pm
i was browsing through the forum, and i noticed there were no replys to your post.
in my experience benzyl alcohol tends to tail dramatically as the system becomes contaminated. if you're calibrating a typical 8270 list of compounds, the tail may actually overlap 2-methylphenol, making integration difficult.
you might try clipping off a loop or two of column, if you think this might be the problem.
on the other hand, i recently had a problem calibrating carbazole (which tails in a similar way), but i had just installed a new column, so no obvious tailing. who knows?
sorry about the late response, but i just joined this forum. you've probably worked this out by now.
in my experience benzyl alcohol tends to tail dramatically as the system becomes contaminated. if you're calibrating a typical 8270 list of compounds, the tail may actually overlap 2-methylphenol, making integration difficult.
you might try clipping off a loop or two of column, if you think this might be the problem.
on the other hand, i recently had a problem calibrating carbazole (which tails in a similar way), but i had just installed a new column, so no obvious tailing. who knows?
sorry about the late response, but i just joined this forum. you've probably worked this out by now.
2 posts
Page 1 of 1
Who is online
In total there are 223 users online :: 1 registered, 0 hidden and 222 guests (based on users active over the past 5 minutes)
Most users ever online was 10230 on Thu Dec 04, 2025 5:56 am
Users browsing this forum: Amazon [Bot] and 222 guests
Most users ever online was 10230 on Thu Dec 04, 2025 5:56 am
Users browsing this forum: Amazon [Bot] and 222 guests
Latest Blog Posts from Separation Science
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science