Calculating unknown concentration by using Millinum32

Discussions about chromatography data systems, LIMS, controllers, computer issues and related topics.

4 posts Page 1 of 1
Dear Sir

I constructed calibration curve is process by creating processing method, also unknown sample chromatogram was processed by using the same method. I got the equation from software by using 1/x as weighing factor y=0.00144X+0.02694. As the calibration plot was plotted between the Area ratio Vs Concentration. Here in this calibration plot equation y is area ratio where X is concentration.

In case of unknown sample the area of analyte is 64152 and Internal STD is 180630. If we want to use the equation (y=0.00144X+0.02694) that means we have to use area ratio of Analyte to IS. The area ratio is 0.3552. If we will put this value in to equation y=0.00144X+0.02694 as ‘y’ value we should get concentration of unknown sample. As we will put this value in this equation the concentration for unknown sample is 277.96ng/ml.

But if will see the chromatogram, by using the processing method (which used for construction of calibration curve) concentration is 11374.264ng/ml. as y=mx+C is worldwide accepted equation to get the concentration of unknown sample by using response.
So why these descrapencies are there? Is there any thing we have to change setting in software as I am using Millinium32 version4.0 software.



If any another method is there to calculate the concentration from software, kindly intimate us.


If any one can solve problem wants the unknown sample chromatogram with Calibration curve kindly provide the e-mail ID as can not attach files here
Thanking you

PRASHANT

Is there a 6mL/250mL or similar dilution you've neglected to tell us about?

If so, you mey be seeing a raw amount in the one screen and the dilution factor being applied to the other number.
Thanks,
DR
Image

Dear Sir
I have constructed the calibration curve in rat plasma so i took the processing plasma volume is 200ul. Sample plasma sample volume i took for animal study sample also. So it is no way related to dilution.

PRASHANT

Are you sure you didn't put any number in the weight / dilution custom field? I think that was the point DR was referring to.
4 posts Page 1 of 1

Who is online

In total there are 26 users online :: 0 registered, 0 hidden and 26 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: No registered users and 26 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry