Dear Uwe
thank you for your response.
Do you remember our mail exchange about a year ago?
I created my own gradient transfer calculator and had asked you about some details.
My calculator is working well (afaik) and I'm using it quiet often for my work. There, I implemented the correction for the delay volume, so this parameter is fine, thanks.
Slowly but more and more, I'm getting my colleagues to be aware of such details.
But, now I "discovered" this system parameter (bandspreading volume), which I think won't hurt to know its value.
Background:
Since we run into some issues when we applied existing methods on our Agilent RRLCs, I would like to know our systems better, so I can fasten the troubleshooting/-solving process or even avoid problems.
What we often get now on the RRLC, is the "volume overload" issue due to too high solvent strength of the sample. Since there were no problems in the past on the other HPLCs, this parameter wasn't realy considered.
But now, the "optimized" RRLC with its short and narrow tubing seems to be very sensitiv to this.
What we conlcuded from my last post (
http://www.sepsci.com/chromforum/viewtopic.php?t=8252) is, that our Waters systems are less senitive to this cause of the injector design.
But we don't have a clear occasion why the third systems (Varian) are less sensitive to this too.
One of my guess is that these systems have more bandspreading volume due to thicker tubing.
(I won't be surprised if we didn't used all the performace of the system and falsely assigned problems to column failure in the past...)
Back to the topic:
I re-evaluated some existing data from the PQ-test (Waters) in the manner of Tom.
I took the width@4.4% and divided it by 5, raised it to the power of 2 and plotted it versus tR^2
Linear fit (n=3) gave a r^2 of 0.9998, N=(1/slope)= 10211, and the extra column volume = 33µL, for our Alliance system.
(Column was a Symmetry C18, 3.5µm, 4.6x75mm)
For the second system, a Waters 600 modular: r^2=0.9997; N=10811; V(ec)=31µL
In addition I carried out some other test suggested on a Waters troubleshooting presentation:
- replaced column with a union
- MeOH 1 ml/min, isocratic
- injected various µL of a solution of Aceton 6 g/L and Acetophenon 60 mg/L in MeOH (gave about 700 mAU with 5µL)
- set the PDA (2996) on single mode to 254 nm, 10 Hz, no filters, no delay
- evaluated the width@4.4% and tR
For the Alliance I get the a system void volume (peak appex) of 100µL, and a bandspreading volume of 54µL.
For the 600er sytem: V(sys)=100µL; V(bs)=63µL
On the 600 the value of 63 staied +/- constant from 1 to 5 µL and increased then for increasing injection volumes.
(If I subtracted the injection volume from the V(bs). the values were constant up to about 10µL, @58µL)
The value of 33µL looks a bit too low for me, while I think the 60µL are plausible and quiet good.
QuestionS:
Are the values of the two methods comparable or are they measuring some different volumes (e.g. volume column-to-PDA vs. inj-to-PDA)?
Or should they be more or less similar and probably my data from the PQ-tests aren't good enough?
Is the correction for the injection volume not made cause it's not relevant or would it be even false to do it?
Thanks a lot!