lifetime of filter line(frit)
Posted: Wed May 07, 2008 2:28 pm
by coquim
Hi people! i want to know what lifetime is considered "normal" for these little filters line, (if we do all correct using procedures like filter and sonicate solvents etc etc.)
Thanks!
Posted: Wed May 07, 2008 3:21 pm
by tom jupille
Depends on your solvent. For pure organic or water, essentially the lifetime of the instrument.
On the other hand, I've experienced situations where we had to clean the frits (boiling in 50% nitric acid) weekly when using aqueous buffers (microorganism growth).
Posted: Wed May 07, 2008 6:21 pm
by JJG
Another factor you should consider when determining frit lifetime is what type of mixing system you have. If your instrument uses a low pressure mixing system with multiple solvent lines, and some of the frits are more dirty than others, then it could lead to irregular mixing by your proportioning valve. For this type of system, it would be good to replace or clean them yearly as preventative maintenance.
On the other hand, if your instrument uses a high pressure mixing system, then dirty frits would have less influence (unless they are really clogged up). Each pump will continue to pump the desired flow rate and the desired amount of solvent from each pump will also be delivered to your column.
Posted: Thu May 08, 2008 2:25 pm
by coquim
Thanks!! both of you, four your prompt reply!
Posted: Thu May 08, 2008 6:59 pm
by randy
Are we talking about the inlet frits that are placed at the end of the eluent lines in the reservoirs, or in-line filters?