No signal HeadSPace GCMS

Posted: Thu Mar 22, 2018 3:14 pm
by Lumi
Hi, can someone help me?
I'm trying to develop a method for residual solvents(Methanol and Ethanol) in a solid sample. I am using HeadSpace Turbo matrix 16 coupled to GC HP6890 5973Mass Selective detector. I was using different methods that I have found online, but I don't get any signal.
MeOH and EtOH standards been made up in Water and 1ml of it been transfered to the Head SPsce vial, I heat it for 20 min at 80 degrees. If i'm running just water blank I get the same result,no differences. To me seems like there might be something wrong in the system, the tune evaluation looks perfect, no leaks.
I have checked the gas and looks alright through HeadSpace transfer line and I was running a standard using just the GCMS to see if there is any blockage in the system and I got poor sensitivity. Can someone advise me what might be the problem?
Thank you
Lumi
MET

Re: No signal HeadSPace GCMS

Posted: Thu Mar 22, 2018 4:44 pm
by Gizmo
Without knowing the details of your method its hard to say but I'm going to throw out a wild guess as I once ran into a similar problem that was fixed by checking the scan range of the MS and making sure it reached down into the ions that ethanol and methanol throw off.

Re: No signal HeadSPace GCMS

Posted: Fri Mar 23, 2018 12:07 pm
by Bigbear
I'm with Gizmo. For those compounds FID would work better.

Re: No signal HeadSPace GCMS

Posted: Fri Mar 23, 2018 12:56 pm
by Lumi
Hi, thank you for your answers
My method has a scan range 25-100amu, I'm only looking for MeOH and EtOH.

Re: No signal HeadSPace GCMS

Posted: Fri Mar 23, 2018 2:55 pm
by Gizmo
Based on a what's been indicated so far I'd start looking at the injector - change the liner, septum, etc. etc. and make sure I got an appropriate response with just a directly injected standard without the headspace system hooked up.

Re: No signal HeadSPace GCMS

Posted: Fri Mar 23, 2018 5:47 pm
by Bigbear
Think you need to scan down to 10.