QC for Glycol Analysis
Posted: Wed Mar 21, 2018 3:11 pm
by sohaib1
Can anybody suggest how to prepare QC samples. I am analyzing Ethylene and Propylene Glycol. Can i prepare QC from Stock Calibration solution or it has to be diffrent. I need help in QC Please. Thank you in advance
Re: QC for Glycol Analysis
Posted: Thu Mar 22, 2018 6:35 pm
by James_Ball
It depends on what agency you are reporting data to.
Most require that a calibration curve be validated by a QC sample that is prepared from a source separate from that of the calibration curve standards, but ongoing QC can be from either source. If you do not have two sources of ethylene and propylene glycols, then make two separate preparations of a stock dilution standard from the original compounds. This will show at least that there is no problems with the preparation.
Posted: Thu Mar 22, 2018 9:40 pm
by sohaib1
So are you suggesting i make separate QC from pure Ethylene Glycol and Propylene Glycol. Like lets say make 50 PPM Ethylene Glycol and 50 PPM Propylene Glycol?
Posted: Fri Mar 23, 2018 6:26 pm
by HPLC chemist
Why not just obtain a previously tested QC sample and spike it with various concentrations of Glycol and EO?
Posted: Fri Mar 23, 2018 7:18 pm
by James_Ball
If you have a pre-made mixture from a supplier then you can use that also. We normally order a solution from one supplier to use for making the calibration curve and another solution from a different supplier to do a QC standard. Restek, AccuStandard, Ultra, or any of the many other suppliers will probably have what you need for about $35.
Posted: Fri Mar 23, 2018 9:50 pm
by sohaib1
OH ok thank you. So what QC will i need to make. The results will be reported to an internal department.