Dear Duantech,
I have used a CAD-Detector for pharmaceutical impurity analysis. We wanted to come up with an assay/purity method for an API.
I found that it was not possible to detect 0,05% impurity in a 100% sample of the API. This was due to the fact that the attenuation/range was not dynamic enough (not to be confused with sensitivity!). This means either I could detect the impurities and the main compound was cut (note the column was not overloaded, it was only the signal was too high fopr the detector), or I could detect the main compund well, but then the noise was too high.
Therefore switching the attenuation during a run would be nescessary.
As for the response factors, this is right for isocratic elution. It is however not right for gradient elution. It would therefore be nescessary to calibrate at different organic modifyer oncentration.
Else, the detector worked very well. It is easy to handle and detects a lot. In our case I could prove that we had substantial amounts of an impurity without chromophor in our sample. This way I was able to close the mass ballance,...
Hope this helps,
Daniel.