Rtx-XLB vs. DB-XLB capillary columns

Discussions about GC-MS, LC-MS, LC-FTIR, and other "coupled" analytical techniques.

3 posts Page 1 of 1
I am working on an application looking at PAH-type hydrocarbons in seawater at ultra-trace concentrations (0.1-10 ppb). I have about 40 compounds to deal with and am planning on using direct immerseion SPME coupled directly to Agilent GC-MSD. I have decided a 25m x 0.2mm id column with 0.33u df is best option. Any expirecince with either/both Rtx-XLB (Restek) or DB-XLB (Agilent/J&W) phases?

djletin,

How hot do you intend to go? You might consider using a SimDis column instead, depending on your application. I can post a part number for one if you need me to.

Best regards.

djletin,

You might also consider the RT-PAH or Rxi-5Sil columns. On Restek's web page is an app note for low level PAH's by MS/
SIM.

Best regards.
3 posts Page 1 of 1

Who is online

In total there are 228 users online :: 2 registered, 0 hidden and 226 guests (based on users active over the past 5 minutes)
Most users ever online was 10230 on Thu Dec 04, 2025 5:56 am

Users browsing this forum: Amazon [Bot], peter36.england and 226 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry