Double Peaks of the Same Analyte
Posted: Mon Apr 28, 2008 11:46 am
by Envirosun
I am running a 6890 Agilent GC with splitless flow. I recently dropped my run time to 14.25 minutes for pesticides and the chromatography looked great. However, suddenly, I began getting what appear to be double peaks for the same analyte: a large normal peak and a small hump attached following the initial peak. I can only assume this second small shoulder is part of the initial "parent" analyte, since it is occurring for each peak in the first half of the run. I switched back to my normal run of 24 minutes to see if there were any change--there was not. Has anyone experienced this, or are there any suggestions in correcting the problem. It is making integration and calibration impossible--I cannot rely on what may be part of the same analyte. Thanks in advance for any advice!
Posted: Mon Apr 28, 2008 9:16 pm
by Rick
What is your old and new oven temperature pgrm?
What is your solvent and column?
Posted: Mon Apr 28, 2008 11:49 pm
by jh1
Which pesticide and have you purchased new standards recently?
This secondary peak could simply be a different isomer of the pesticide.
Posted: Fri Aug 01, 2008 12:44 pm
by fsistere
Hi
I'm sorry about my (very) bad english :(
If appears doublepic in many pics of your cromatogram it's possible a problem with the injection.
Posted: Fri Aug 01, 2008 2:38 pm
by s2008
probably the column was dead or overloading. Check with new column and smaller injection volume.
Posted: Mon Aug 04, 2008 11:41 am
by danicrd
try lower initial temp of column: the problem is simply injection problem
good luck
Posted: Mon Aug 04, 2008 2:34 pm
by Suresh Seethapathy
I have seen this with a Thermo Focus GC and AS 3000 autosampler. Simply increasing the injection needle dwell time to 2 seconds solved my problem. Never noticed anything like that with the Agilent autosampler which seems to have the fastest injection i know of.
Posted: Tue Aug 05, 2008 11:24 am
by frage
Hello!
test see if the gold Plated is clean?
P/N 18740-20885