Waters 2998 ethernet communication issue
Posted: Thu Mar 08, 2018 5:45 pm
by Biofix
Hello Everyone,
I'm (a rookie) technician in an analytical laboratory (Nice Area, South France, apology for my poor english)
It's been a few days since we tried to put a Waters AutoPurification System back into service (improperly stored since 23 months), without success.
Here's our issue:
All modules detectors seems to properly start up except one : 2998 PDA (with new lamp) detector.
Although the front instrument lights are steady green, the pda is not recognize by MassLynx software (main soft, and other utilities soft like Autoloader & Co...)
A quick look behind the instrument, onto the ethernet hub, shows there's no communication between PDA detector and SFO module (and consequently MassLynx PC, thank you capt'n obvious).
Usual checks (ethernet cables working / all hub positions working) do not solve the problem.
Waters support team, for the moment, not succeed to diagnose.
So if anyone already experimented that problem, or could suggest a point that we have not properly check, it would be valuable.
Once again apology for my english
Re: Waters 2998 ethernet communication issue
Posted: Thu Mar 08, 2018 10:09 pm
by Hollow
Hi
did your usual checks also include a check for the IP/Subnet addresses, that means can you ping the IP of the PDA?
If not, check the IP settings of the PDA and the network interface card.
Maybe ping one module at a time, in case two have the same IP address.
Maybe also check the ARP table (and/or clear it) to see which MAC addresses are recognized.
For Empower (guess will be similar for Masslynx) normaly the systems are in the 192.168.0.X and guess something like 64.1.1.1 networks (maybe I can check it next time I'm in the lab).
Unlikely but also have a look at the firewall settings (if any). Maybe a certain IP address is blocked.
In case you connect the modules directly to the PC (without the switch in between), it's highly possible that you'll need a crossover adapter/cable. Maybe check with a known good one, if you still can ping this one.
UPDATE/Correction:
the Networks are: 192.168.0.1 / 255.255.255.0 and 64.1.1.1 / 255.0.0.0
Posted: Sun Jan 12, 2025 10:15 am
by fareedtayyeb
Well I have replaced new lamp and new flow cell but 2998 never eve passes wavelength Calibration test neither the Erbium wavelength values gets closer to required wavelength ranges as mentioned in OP guide..
Flow cell is firmly slot in with guide pin also D2 lamp is properly installed.
On startup green lights comes normally then after few moments and erbium.filter motor movement it appears with red light with D2 lamp still glowing and give Wavelength Calibration Error
Tried almost all remidys changing solvents from water to formic acid,IPA,ACN etc.
Re installed the Firmware and ICS
I WISH TO KNOW ON STARTUP WITHOUT ANY ETHERNET CONNECTIONS WHY THE DETECTOR GIVE THESE ISSUES ONCE WE HAVE NEW FLOW CELL AND LAMP.
ALSO WHAT SHOULD BE THE READ ENERGY TYPICAL VALUES ON ERBIUM AND CLOSE POSITIONS IN 2998 CANSOLE AT 254NM OR 190-500 OR 190-800 NM SETTINGS ITS NOT MENTION ANYWHERE IN 2998 OP OR ANYOTHER LITERATURE AVAILABLE FROM WATERS
PLZ GUIDE WE HAVE COUPLE OF UNITS SITTING SICK WITH SAME PROBLEMS
REGARDZ
Posted: Sun Jan 12, 2025 10:40 am
by fareedtayyeb
REGARDZ