C4 Column Suggestion to analyse a 4kD(peptide) 60kD(protein)

Posted: Fri Mar 02, 2018 3:28 am
by Faith
I need to quantitate a ~3.5 kD peptide with 2xC16 chains attached, and a 55 kD protein in the supernatant. The compounds are in a DPBS pH7.2 buffer.
It needs to be a very sensitive method (<<50ug/mL).


I have had issues in the past and had developed two separate HPLC methods (using SEC for the protein) and a C8 100A for the peptide.

I only had a C8 column on hand at the time.

But, I was wondering if I could analyse them both on a C4 column, possibly in the one method.
However, would I need a 300A pore size column because of the protein? Would this mean the peptide would take ages to elute? I thought a C4 column would be good because I wouldn't require as much acetonitrile to elute the protein, and hence it would not crash out.

I have also learnt in order to keep the peptide in solution (because of the attached lipids) to heat the auto-sampler and the oven compartment to 37C.

I was wondering if there was an ideal ~2um; 100x2.1; ??Angstroms; perhaps coreshell technology UPLC column available. A require a very sensitive column. I have attached a high sensitive flow cell to the Shimadzu DAD.

I would very much appreciate any suggestions. Protein and peptide analysis is relatively new to me.

Re: C4 Column Suggestion to analyse a 4kD(peptide) 60kD(protein)

Posted: Fri Mar 02, 2018 3:42 am
by mattmullaney
Edited:

Hi Faith,

I have only a small amount of experience with protein separations, and these are new products; however, they deserve consideration in my opinion:

http://www.mac-mod.com/products-halo-bioclass.php

The HALO 400 columns may serve you well.

Re: C4 Column Suggestion to analyse a 4kD(peptide) 60kD(protein)

Posted: Fri Mar 02, 2018 11:14 am
by HPLC chemist
Does anyone even make C4 columns anymore? Brownlee used to make them. I would try your C8 column particularly if it is short (10 - 15 cm long and can perform gradient). There is newer technology out there!

Re: C4 Column Suggestion to analyse a 4kD(peptide) 60kD(protein)

Posted: Fri Mar 02, 2018 11:47 am
by Faith
Thanks MattM

I clicked your link. Read all about the columns.

Thought the HALO 400 Protein C4 column looked perfect. As the two compounds I want to analyse in the one run are ~3.5kD and ~55kD. And..Included in the info. on the website was a chromatogram with 13.7KD (Ribonuclease) and 82KD (enolase) in the one run on this particular HALO column. The Rib A came out just after 2 mins, but the gradient started at 25% acetonitrile, so there is a possibility the 3.5 kD may be retained.

I wrote to the distributors of the column in Australia with a few questions, and am waiting for them to respond. But the column seems ideal. Very sensitive.

Thankyou so much.

Re: C4 Column Suggestion to analyse a 4kD(peptide) 60kD(protein)

Posted: Fri Mar 02, 2018 12:02 pm
by Faith
HPLC Chemist,

Yes they still make C4 columns. But your right, perhaps I should just ring up some companies and ask them to suggest something with all the new technology now.

I will also try the protein on the C8, but I am worried that I will have to use too much acetonitrile, as I think I went up to 52% to get the peptide off the column, and I don't want the protein to crash out having to use a lot more.

Re: C4 Column Suggestion to analyse a 4kD(peptide) 60kD(protein)

Posted: Fri Mar 09, 2018 10:24 pm
by mattmullaney
Hi Faith,

Just curious, which way did you decide to go for the stationary phase?

Re: C4 Column Suggestion to analyse a 4kD(peptide) 60kD(protein)

Posted: Fri Mar 09, 2018 10:55 pm
by Faith
I have ended up ordering
Halo 400 Protein C4, 2.1 x 100 mm, 3.4 um column.
Am very excited, coreshell technology and the range is from 2 kD to 500kD. They have a chromatograph with a number of compounds in it on the site (from 13.8 kD (Rib A) to 88kD (enolase) ) if I remember correctly. But, that gradient started at 24% acetonitrile, and RT was just after 2 mins for the 13.8 kD so feel confident I can retain a 4kD peptide. Then I later read, the range of the column starts at 2kD. I have to wait three weeks for the column to arrive. :cry:

Re: C4 Column Suggestion to analyse a 4kD(peptide) 60kD(protein)

Posted: Sun Mar 11, 2018 11:59 am
by mattmullaney
Best Wishes Faith!