running hp1100 without every detector on.

Posted: Mon Apr 21, 2008 6:43 pm
by matthew
Hi, I have a silly question involving Chemstation and our old HP1100. I have an FLD and a PDA on the unit, and it seems that I have to turn on both detectors to run a sample, regardless of my needs.

I'd like to save the lamp in either unit, but how do I get the lovely 'ready' light when a unit is off? Do I have to take a unit out of the configuration of Chemstation (A.10)?

Somewhat separate question... should I upgrade the software to B.0.3?

Re: running hp1100 without every detector on.

Posted: Mon Apr 21, 2008 8:19 pm
by zokitano
Hi, I have a silly question involving Chemstation and our old HP1100. I have an FLD and a PDA on the unit, and it seems that I have to turn on both detectors to run a sample, regardless of my needs.

I'd like to save the lamp in either unit, but how do I get the lovely 'ready' light when a unit is off? Do I have to take a unit out of the configuration of Chemstation (A.10)?
Dear Matthew,

You can solve your problem easily. In your LC method (when configuring the method parameters> you could try this through the option in Chemstation: Edit entire method) when you'll get to the detector parameters, you could exclude the desired detector from your method. By this the loaded method won't search the excluded detector so you could switch it off (and save detector lamp(s)).

That can be done, for example if you want to exclude the PDA detector from your method, by switching the lamps (UV and visible) "off" in the "control" window under the detector parameters and also choose in the "signal" window (under detector parameters) spectra not to be stored.
In that way, the method won't search the excluded detector, you could switch it off, and then chemstation will give you the ready status (of course if the method included instrument modules are already turned on and initialized :wink: )

Hope this helps

Regards

Posted: Mon Apr 21, 2008 8:53 pm
by matthew
thanks zokitano!
My real difficulty was turning the FLD off, but indeed, if one clicks:

1) full >>
2) special setpoints

on the FLD dialog, you can then click:

'enable analysis when lamp is off'.

Finally! Thanks so much.

Posted: Tue Apr 22, 2008 2:56 pm
by Consumer Products Guy
Am I missing something, like why couldn't you remove the detector from the Configuration and just power off that module? Like with ChemStation Configure Editor (after exiting ChemStation) or on Chemstation Menu under Instrument, Configure 1100 Access...

Posted: Mon Apr 28, 2008 5:08 pm
by matthew
Sorry that I didn't see your reply earlier, CPG. I had considered removing the FLD from the configuration (above). It's more or less the same, but to be greener, perhaps I should just remove the thing from the configuration.