HPLC quantitation formula doubt
Posted: Sun Apr 20, 2008 4:13 pm
by vellaikumar
Dear Members,
I have one doubt on HPLC quantitation formula,
Amount in micro g / g (ppm) = As/Astd X M/M1 X V/V1 X F
where,
As= sample area
Astd= std.area
M= std injected (micro lit )
M1= wt.of the sample (g)
V= vol of final extract (ml)
V1= sample injected (micro lit)
As per the calculation, the final result accounts for ml/g and
not for micro g/g.
Can any one explain this.
Posted: Sun Apr 20, 2008 8:50 pm
by tom jupille
Where did you get that formula?
You have M and M1 flipped, but it's very confusing to use the same primary symbol (M) for both a mass and a volume (M1).
Look at it this way (I'll refer to your standard as "1" and sample as "2"):
for equal injection volumes, the concentration of sample is equal to the area ratio times the concentration of standard:
M2/V2 = (A2/A1)*(M1/V1)
where
M1 is mass of standard
V1 is volume of standard solution
M2 is mass of sample
V2 is volume of sample solution
If you inject different volumes of sample and standard, you have to correct that by the inverse of the injection volume ratio (we'll call injection volume "I"). That makes the final formula:
M2/V2 = (A2/A1)*(I1/I2)*(M1/V1)
where
I1 is injection volume of standard
I2 is injection volume of sample
the final result (M2/V2) will have the same units you used for M1 and V1 (e.g., g/mL, g/L, μg/mL, . . . , whatever). If you want ppm, you will also need to use either μg/mL or mg/L and then factor in the specific gravity of the solvent).
Posted: Wed Apr 23, 2008 8:45 am
by vellaikumar
Dear Tom,
Thanks for your reply. Can u please explain me in practical.
by vellaikumar
Dear Tom,
Posted: Wed Apr 23, 2008 12:52 pm
by AdrianF
If you want to understand how to work it out I suggest you forget about the formula and work it out from first principles.
The principle is that the area is proportional to the amount injected.
1. From a knowledge of the concentration work out the amount you injected in the standard
eg 100ug gives an area of 1000
therefore 1ug will give an area of 1000/100 = 10
2. From the area of the sample work out the amount in the sample
Say your sample has an area of 700 the amount in the sample is 700/10 = 70ug.
3. Calculate the amount in your original sample
Lets say the volume of the injection is 20ul and the sample of 1 tablet was dissolved in was 100ml
Amount in the flask is 100x1000(to convert to ul)/20 = number of times 20ul can be fitted into one flask. = 5000
amount in flask is 5000 x 70ug = 350,000ug
= 350,000/1000 mg = 350mg
I think it is important to think through your calculations from scratch. By doing so I have found a number of errors in methods I have been given.
Sorry if this seems very basic - hope its helpful